Tension calibration - possible to add cal points? ? (and is 2 point good enough)

Started by mawrick, August 06, 2014, 03:14:54 AM

Previous topic - Next topic

mawrick

What experience does people have when it comes to the two point calibration of tension that's the default in Warrior ?.

Also I'm wondering if it's possible to add more calibration points to the normal way one does the tension calibration.


For us it seem to be somewhat "off" at times - but I'm not sure if this is due to the 2 point calibration or not.


alioda

While making your calibration check if the raw value is not saturated at 5v as if so, try to reduce the warrior input voltage from your tension system to 5v or less then two point calibration will be Ok.