Started by RK, November 05, 2009, 03:17:00 PM

Previous topic - Next topic

RK

After importing a working service and tools from one computer to the next, I get Sensor syntax DSP tel 1/2/3/4, sensor cant initialize, not measuring tool voltage or current when loading it in the aquisition.  What is the problem?
RK

adamg

The most likely reason is that the dsp script file is missing (it is not currently part of the export/import process).

Looking at the service with Service Editor, one of the lines in the Device box will say DeviceN=DSP or SDSDSP, script="???????"

The ????? will be the script file that the service needs to be fully configured. You will find this file on the computer that you exported the service from in the \warrior\bin32 folder. Just copy that over to the new computer's bin32 and the service should load as normal.